    Experiment to View T1 Clock Recovery

     

     

    Introduction

     

    This experiment demonstrates how a T1 data communications code can be viewed by the ADC200/20 oscilloscope and how a simple circuit can alter the code to recover the embedded clock signal. The figure below illustrates various baseband data codes and their respective frequency spectra. The T1 signal is the Bipolar AMI RZ.

    image showing the waveforms for communication data

    This experiment demonstrates:

    • What a very popular serial data communications signal looks like.
    • What the frequency spectrum of that signal looks like.
    • A simple circuit to extract the embedded clock signal in the data for signal regeneration purposes.

    This experiment is suitable for:

    • College or University Data Communications Course.
    • Data Communications Professionals Carrier/ISP.

     

     

    Equipment required:

    • ADC200/20 oscilloscope running PicoScope oscilloscope software.
    • T1 Test Generator. Wandel & Goltermann DSA15.
    • Coaxial cable with BNC T Connector Terminated in 100 ohms.
    • 2 x Alligator clips
    • Diode Bridge (4 diodes soldered in bridge configuration 1N914)

     

     

    Safety

    Always keep in mind that the ground of the ADC200/20 is common to the ground of the PC and thus common to AC ground.

     

     

    Experiment set up

    The set up is shown in the sketch and picture below:

    connection of the ADC-200 to the test set

    Figure 1: experiment set up

     

     

    Experiment procedure

     

     

    1. T1 Test Set Direct to ADC200/20

    Using the correct bantam test cord, connect the T1 test set output directly to the Ch1 input on the ADC200/20. A coaxial cable with alligator clips can be used to connect to the tip and ring on the balanced bantam cable. Ensure that the Ch1 input is terminated in the correct T1 impedance of 100 ohms.

    Set the T1 Test set for the standard Bell QRSS test waveform. This generates a predetermined stream of random bits. This waveform can be easily captured by the ADC200/20 in single shot mode. The capture is shown below:

    random data captured by PicoScope from the ADC-200

    The spectrum of the QRSS signal is shown in the first figure. It is a smooth hump with very little energy at DC, peak at 772 kHz, and null at the bit rate of 1.544 MHz. The spectral capture on the ADC200/20 is shown below.

    captured frequency spectrum using PicoScope

     

     

    2. T1 Clock Recovery

    In order for a digital signal to be useful, it must be periodically regenerated to remove unwanted noise. To do this, a clock signal must be present at the data rate for sampling purposes. In order to do this, the T1 signal can simply be put into a full wave diode bridge. 4 ordinary diodes like 1N914 diodes can be used for this purpose as shown below.

    full wave rectifier

    Connect the T1 test set with alligator clips so that the signal is at the AC points of the bridge. The ADC200/20 tip will connect to the positive of the bridge and the ring to the negative.

    The T1 signal is now rectified and the capture is shown below.

    captured screenshot of rectfied signal using PicoScope

    The spectrum of the rectified T1 signal now shows a strong recovered clock at the bit rate of 1.544 MHz!

    spectrum of rectified signal using PicoScope

     

     

     

    Credits, comments and further info

    This experiment was written by Jeremy Clark and Kyle James Mcneil at Seneca College, Canada.